Strange artifact in DWI - differential slice-wise scaling in AP directions

Dear experts,

I am writing this post in the hope that you can help me to correct for following artifact I encounter in the data of one of our studies. This artifact is manifesting by differential slice-wise scaling in AP direction (AP direction is phase-encoding direction). The scheme is 2 slices stretched followed by 2 slices squeezed an so on. The artifact manifests itself only in some specific diffusion sensitization directions (in about 10 from 64 directions) and AFAICS it seems that it is present in all of the subjects in the study (consisting of approx 200 subjects). Some of directions are not affected in all subjects, but there are several specific directions which are affected in all subjects.
It is standard Siemens Trio single-band 64 direction sequence, b=1100. Standard interleaved acquisition (that is why it is puzzling me since in the artifact there is interleave in slice scaling in period of 2 slices, not in one slice).
See the screenshot demonstrating the effect:
DWI vol with artifact:


subjequent DWI vol without artifact:

I hoped that the new slice-to-vol registration in eddy could be able to correct this artifact (Jesper expressed hope that it could be the case in the personal communication) but according to my tests with the new eddy it is not the case, the artifact is present even in the eddy-corrected data.
It is probably not the movement problem since the slices are not moved to each other, there is systematic scale between them.
Do you have any idea of the source of the artifact and how to correct for that?
I think it could be doable to correct it since the artifact is so systematic.
I was wondering whether I could do following (probably not ideal approach, but it is better than nothing):

  1. split affected volume into two separate vols with stretched and squeezed slices
  2. register two vols together by restricting to movement and scale to AP direction (it is possible to do it via mrregister?)
  3. concatenate mutually correctly scaled volumes again together

Do you have better idea?
Any comments/suggestions will be greatly appreciated.
Regards,

Antonin

Wow, I’ve never seen anything like this… If this is a product sequence, and it’s running in traditional single-band mode, I would definitely chase it up with Siemens - it might indicate an issue with your scanner, or a bug in the sequence code that they’ll want to fix ASAP…

Is there any consistency to the directions affected? For instance do they tend to cluster around the PE direction? But even then, it’s hard to envisage any hardware-related issue that could cause this, it looks like a bug in the acquisition or reconstruction code…

As to how to fix this, your suggestions sound like a good starting point, I’m not sure I can add much. The one thing I’d change is maybe trying to avoid registration altogether, and see whether the scale factor that you need to apply is actually reproducible across slices and subjects - in which case you might be able to use mrtransform with a single affine matrix, and avoid errors introduced by registration (you might need to remove or reset the image’s transform matrix for that work if you have oblique acquisitions - and then put it back in afterwards). I’d personally be concerned about having up to scale up the affected slices, it’s not unlikely that the required interpolation will introduce some systematic bias… If the number of volumes affected is not too large, I’d be tempted to discard them. But that might be problematic if the affected DW directions are all similar, since that will likely introduce an orientation bias… Not a simple problem to fix!

Hello I am happy this issue show up here
This is (I think) something we have seen on multiband sequence. It shows up also on b0 volume when they are just after a diffusion.

We report it on cmrr multiband github issues
uncoherent slices distortion · Issue #199 · CMRR-C2P/MB · GitHub

After looking at gradient cable heating it ends up to be due to transient eddy current effect (from the previous diffusion encoding) .

Working with Julien Sein in Marseille, he reproduce the same artefact and I am surprise it is not more known

Julien had a close look to the literature and gives me some references :

C. Hardware correction
Using Siemens service hardware, we map the eddy current along the time axis (Figure 3A). We found the typical compensation of eddy current mainly focused on the time range of 2-100ms, but the amplitude of eddy current around 1-4 seconds is still large. This time range is close to the TR (3222 ms) used in the Lifespan protocol. The super short TR and super high diffusion gradient strength in this MB protocol make it possible that the eddy current affects different images across TR. That is why the b0 image in the Lifespan protocol displays this artifact.
After our re-tuning of the compensation circuits, the eddy current after the specified time range (1- 4s) is also reduced (Figure 3B). Therefore this artifact is deducted significantly in the new images (Figure 4B).

thanks to julien again
http://indexsmart.mirasmart.com/ISMRM2018/PDFfiles/3194.html

For the retrospective correction (I am also very interested in a good solution)
slice to volume correction with eddy does not correct it (I guess this is because they only search for affine registration and we have non linear distortion …)

talking with Sotiropoulos I understood they are working on it with eddy so may be in a near feature …

I am surprise it is not more widespread (may be people do not see it … ?) but clearly it is protocol dependent

Romain,
thank you for sharing the information.

The artifact I was reporting is not from mubltiband sequence, it is standard 64 direction sequence.

It is not correctable by eddy even with slice-to-volume registration as it stands, since AFAIK the slice-to-volume movement is only modeled as rigid-body (translation, rotation).

The effect was to large extent consistent in the directions across subjects. Therefore, we finally did arbitrary differential slice-wise transformation (shrink/stretch) for specific volumes across subjects (to optically get rid of “jags” in sagittal view) + jacobian modulation, and individual exclusion of volumes where this effect was not so systematic.

Antonin

Hey everyone,

we also had this problem for a while and collected a bunch of data with this artefact.

After an in-depth investigation by Siemens it turns out that it is due to eddy currents indeed, depending on:

  • the type of magnet (e.g. the Prisma is more affected than the Vida, since it hat more cold metal)

  • the temperature stability of the magnet (we had problems with the cooling, e.g. due to ice, which made it much worse)

  • the Eddy current compensation calibration (which is affected by the temperature stability of the magnet)

  • the DWI sequence type (it gets worse with high B-values, short TR, large echo spacing, no GRAPPA, basically anything that makes EPI-AP-artefacts worse…)

Since we have all this data with alternatingly shifted slices in AP direction - does anyone have a good clue on how to correct for this yet?

Thank you!!
Marco